1345141993 发表于 2012-12-17 21:01

改用DC的可操作性值得思考,出厂检测打直流太危险了,看能否改设计吧!

silvermin 发表于 2013-1-7 13:46

各位老师说的很对,我摘录出三版8.8.3的正文以及基本原理,希望对楼主有所帮助,谢谢。

8.8.3a)
Alternatively, a d.c. test voltage equal to the peak value of the a.c. test voltage may be
used.

Subclause 8.8.3 a)
The test voltage can be provided by a transformer, by a d.c. power source or by using the
transformer(s) of the ME EQUIPMENT. In the last case, to prevent overheating, the test voltage
can have a frequency that is higher than the RATED frequency of the ME EQUIPMENT.
页: 1 [2]