DSH 419 Internal battery for back-up power
http://decisions.iecee.org/icons/ecblank.gif DSH 419
Internal battery for back-up power
6.8.2, 6.8.2 e)
60601-1(ed.2);am1;am2
Standard(s)- (year and edition):
IEC 60601-1:1988 Ed.2 Am1+Am2
Sub clause(s): 6.8.2 6.8.2 e)
Sheet n°: DSH-419
Subject: Internal battery for back-up power
Key words: Internal battery, Protective earth terminal
Ce confirmed by CTL at its 39th meeting, in Cologne
Question:
The requirement in this clause seems to be a little unreasonable. Assume for example,
EQUIPMENT with an internal battery intended for back-up power in the event of a failure of the
SUPPLY MAINS as a power source. Further, how can one say whether the external
PROTECTIVE EARTH CONDUCTOR arrangement is in doubt or not?
Decision:
The Instructions for use shall include for CLASS I EQUIPMENT the following statement:
“WARNING: THIS EQUIPMENT MUST ONLY BE CONNECTED TO A SUPPLY MAINS
WITH PROTECTIVE EARTH.”