+ c9 S5 p6 a5 u( ]Standard(s)- (year and edition):5 L6 C4 e C0 z. K( |# U
IEC 60601-1:1988 Ed.2 Am1+Am2/ M8 }0 W3 E. @$ p1 g5 b: E
Sub clause(s): 18f)$ I0 P5 N G; U/ j* C
Sheet n°: DSH-432
5 G' p6 L. W& ySubject: Protective Impedance versus noload voltage
1 [1 |/ ^$ F& t4 MKey words: Protective, impedance, earth, connection, no-load
. _/ T* I( v5 P( V/ g; g- J9 fConfirmed by CTL at its 39th meeting, in Cologne
* Y( q1 Z$ Q' p6 H5 \Question:
2 F2 }$ [7 O1 X. J; LEQUIPMENT for example having a rated current of 30 A require a test current of 1.5 times 30 A.$ h L" ^% J. U7 K; x: h$ @9 G
With a maximum impedance of 0.2 W the voltage drop has to be 9 V (0.2 W * 45 A = 9 V). This is+ l/ q+ A# j+ J! N
in contradiction to the required no-load voltage of 6 V maximum.# D# C" M7 d4 ^
Decision:- w5 r x( N' V0 p K( T* l W2 Q
Measuring the protective earth connection has in fact two reasons. It is to determine impedance) E. l9 k/ o( W% b$ I0 G
and cross-sectional area of protective earth connections. For a measuring current of 25 A both
1 R: `2 B5 F8 O) U% v) G; ycan be done with one measurement. Requiring a measuring current of more than 25 A it shall be
! l) p; v! I, Q$ J- g6 [splitted up into two measurements. In this case, the impedance shall be determined first, using a
' @2 u J+ o( _% I2 Z& fvoltage not exceeding 6 V.
. d% @7 r2 P. U" eIf cross-sectional area of the protective earth connections cannot be determined as equal to the7 x4 n+ l' }/ a1 C7 D
one for the phase by measurement of the area, then measurement with current shall be from a" [. N, J/ t! I. A3 B2 u; }# q: e
source with a higher voltage than 6 V.