灯管用第二种方法测试电压24.2V,算出来有48.4mA。而我在实验室用228 表测试只有2.5mA. 不知原因出在哪里?现在负责的工程师坚持第二种方法的测试结果,因为标准根本没提到可以用228表测试。
7 {% |4 ~& U) M( S1 U$ T# Z
; |' {9 h1 a9 q! K% a/ K% C4 t! |( a4 E4 N% e2 y6 t
第一种方法:Readings were taken directly in MIU using a suitable leakage current meter (i.e.: Simpson 228 set to "let go", etc.).
5 P/ Q/ X+ B% a
6 @8 y9 |1 S; W1 I Q$ C( D第二种方法:Readings were taken in millivolts using a suitable voltmeter connected across the test circuit defined in UL935, Figure 24.2. These readings were then converted to MIU by dividing them by 500.